Chris Rasmussen
Film Handler

Posts: 7
From: Berkeley, CA, United States
Registered: Feb 2005


 - posted 06-28-2008 02:38 AM      Profile for Chris Rasmussen   Email Chris Rasmussen   Send New Private Message       Edit/Delete Post 
I’m trying to determine the cause of base-side, horizontal scratches, running from beginning to end of two prints in a tri-plex.

The scratches are razor thin, irregular and densely packed – around ten to fifteen per four-perf frame – and run fully across the surface, past the sprocket holes and out to the edges. The scratches vary, some not fully stretching across the film’s width, but with no regularity; some scratches run from the soundtrack side inwards, some from the other and some just in the center. There is no discernable pattern. The scratches are consistently random, with a full gamut appearing several times per frame.

The equipment in both houses wherein this may have originated: Century heads (one JJ and one SA) and TR4 soundheads upgraded with red LEDs. Both have Component Engineering automation/failsafes. The platters are AW3 with newer style green rollers, and both platters are within five feet of the projectors.

The film transport paths have been thoroughly inspected: all rollers are properly aligned and turn freely. Pad rollers turn freely, have correct tension and proper alignment. Film gate tension and alignment are likewise correct. AW3 platter return arms have proper tension and all platters are properly timed. The slide bar orientation pins are all correctly aligned and the roller cluster rest plates haven’t been deformed, so the clusters sit at the correct levels to feed film onto the platters without incurring scratches.

The circumstances: the opening projectionist noticed the scratches on one print at startup of the first matinee show on Thursday. I was there, inspected the film feeding from the platter and noted that the scratches had occurred previously. As it turns out, the prints in two sides, both of which had played scratch-free for several days, seemed to have been damaged on Wednesday evening. Unfortunately, a print swap between the two booths leaves me unable to determine in which booth the problem struck.

The conundrum: there is no obvious physical cause for the scratches. There are no accompanying vertical scratches. The scratches are perfectly horizontal, not diagonal, like platter-feeding scratches from improperly set roller-cluster height. The prints do not seem to wind overly tightly on platters in either booth, which I mention because it’s been suggested that excessive take-up platter tension combined with a hesitant roller might cause the scratches. I don’t believe this is the case because the current return arm springs that control the tension have been in place for over a year and are unlikely to get tighter with age.

Has anyone out there seen or heard of a case of perfectly horizontal scratches as I’ve described? I must admit I’m at wits end trying to come up with a plausible cause. I'm a full-time field tech with too many years experience to be thrown by some wimpy little scratches... or so I thought!

With hope and appreciation ~

Chris [beer]

 |  IP: Logged

Brad Miller
Administrator

Posts: 17775
From: Plano, TX (36.2 miles NW of Rockwall)
Registered: May 99


 - posted 06-28-2008 02:44 AM      Profile for Brad Miller   Author's Homepage   Email Brad Miller       Edit/Delete Post 
I would have to blame this on piracy. Clearly some bootlegger broke into your building after hours, stole the print, transferred it to video (scratching it in the process), then put it back before you opened the next day.

If for some oddball chance that actually didn't happen, pictures of the scratches as well as your equipment setup would help get feedback. Even though you gave a great description, I can't think of anything offhand. [Wink]

 |  IP: Logged

Chris Rasmussen
Film Handler

Posts: 7
From: Berkeley, CA, United States
Registered: Feb 2005


 - posted 06-28-2008 03:48 AM      Profile for Chris Rasmussen   Email Chris Rasmussen   Send New Private Message       Edit/Delete Post 
Thanks Brad,

Methinks you may be onto something! Gremlins, while the usual suspects in such cases, have vehemently denied any involvement, so overnight break-in film pirates are likely the only clearly plausible culprits. (Color me in awe of your powers of insight; your Kung Fu is truly more powerful than mine...)

Humbly,

Chris ~

PS - I'll try shooting some good close up pics so you can see the pirate's fingerprints.

 |  IP: Logged

Michael Cunningham
Expert Film Handler

Posts: 186
From: Anchorage, AK
Registered: Nov 1999


 - posted 06-28-2008 06:23 AM      Profile for Michael Cunningham   Email Michael Cunningham   Send New Private Message       Edit/Delete Post 
So they were fine for a while and then horribly scratched in one night? Has the damage gotten any worse since? If not, that's got to be threading error. Not being too familiar with those heads, could this be due to an oversized loop slapping against something? That's the one thing that comes to mind for horizontal scratching.

 |  IP: Logged

Louis Bornwasser
Film God

Posts: 4441
From: prospect ky usa
Registered: Mar 2005


 - posted 06-28-2008 07:09 AM      Profile for Louis Bornwasser   Author's Homepage   Email Louis Bornwasser   Send New Private Message       Edit/Delete Post 
Base side scratches (on a Century) are almost always related to the sound drum. It is the only thing that touches the base side. I would suggest that an occasionally loose threading of the dancer assembly would look just this way. (Also true of Dolby penthouses.)

Easy enough to hold the "damage" up to the drum. Notice the drum does not cover the entire width of the film? Is the damage full-width? Or does it stop to agree with the touching area of the drum? Louis

 |  IP: Logged

Randy Stankey
Film God

Posts: 6539
From: Erie, Pennsylvania
Registered: Jun 99


 - posted 06-28-2008 07:57 AM      Profile for Randy Stankey   Email Randy Stankey   Send New Private Message       Edit/Delete Post 
Did somebody make the loops too big so that the film slaps against the projector bulkhead or some other stationary part?

Just guessing...

 |  IP: Logged

Darryl Spicer
Film God

Posts: 3250
From: Lexington, KY, USA
Registered: Dec 2000


 - posted 06-28-2008 12:31 PM      Profile for Darryl Spicer     Send New Private Message       Edit/Delete Post 
quote: Chris Rasmussen
I’m trying to determine the cause of base-side, horizontal scratches, running from beginning to end of two prints in a tri-plex.

ummm the clue here is horizontal. Sound drum normaly causes vertical base side scratches. Film running across a platter deck with the bar out of alighnment. Missing the final roller on the return to the platter. Missing the first roller after coming out of the feed control roller cluster. All causing the film to drag accross the spinning decks will cause these scratches.

 |  IP: Logged

Carl Martin
Phenomenal Film Handler

Posts: 1424
From: Oakland, CA, USA
Registered: Feb 2002


 - posted 06-28-2008 02:06 PM      Profile for Carl Martin   Author's Homepage   Email Carl Martin   Send New Private Message       Edit/Delete Post 
after sleeping on this (it's my theater), i'm leaning towards darryl's theory. perhaps the films took up on the bottom platters with the takeup cluster at the very bottom of the slider bar.

platter scratches would normally be a bit slanted, but these seem to be consistently perfectly horizontal throughout. can it be that the slowing of the platter over the course of the film is exactly compensated for by the changing angle of the platter edge relative to the film to keep the scratching horizontal? i'll take a closer look to see if there's any hint of a "smile" to them.

 |  IP: Logged

Monte L Fullmer
Film God

Posts: 8367
From: Nampa, Idaho, USA
Registered: Nov 2004


 - posted 06-29-2008 12:30 AM      Profile for Monte L Fullmer   Email Monte L Fullmer   Send New Private Message       Edit/Delete Post 
quote: Darryl Spicer
Sound drum normaly causes vertical base side scratches
..unless the film was wound too tight around the drum due to the lateral guide rollers of someone not knowing the "line and arrow" marking on the roller arms...which I've seen TONS of operations having the film so tight across the scanner drum on both soundhead and Dolby readers...and with "pressing" in horizontal scratches on the base side of the film.

And I've seen the same with platter rollers on the rewind side of the transports-anything when film is under tension.

 |  IP: Logged

Carl Martin
Phenomenal Film Handler

Posts: 1424
From: Oakland, CA, USA
Registered: Feb 2002


 - posted 06-29-2008 02:42 AM      Profile for Carl Martin   Author's Homepage   Email Carl Martin   Send New Private Message       Edit/Delete Post 
my girlfriend's iphone was not up to the task of documenting this scratching. none of the screenshots came out, and this was the clearest shot of the film itself:

 -

i was able to duplicate the scratching by taking up on the bottom platter with the roller cluster at the lowest (unintended) position, as described above. the scratches do in fact stay horizontal (bowed just slightly downward) over the course of a normal-length movie. so that about solves that.

thanks!

 |  IP: Logged

Chris Rasmussen
Film Handler

Posts: 7
From: Berkeley, CA, United States
Registered: Feb 2005


 - posted 06-29-2008 03:08 AM      Profile for Chris Rasmussen   Email Chris Rasmussen   Send New Private Message       Edit/Delete Post 
Many thanks to all for supplying good theories and to Carl for his extensive testing and confirmation that the culprit was indeed pilot error rather than the suspected film pirates or Gremlins.

Cheers!

Chris ~

 |  IP: Logged

Brad Miller
Administrator

Posts: 17775
From: Plano, TX (36.2 miles NW of Rockwall)
Registered: May 99


 - posted 06-29-2008 06:24 AM      Profile for Brad Miller   Author's Homepage   Email Brad Miller       Edit/Delete Post 
Sounds like you have the old style AW3 takeup slidebars. You can either buy a new one (the new ones have a really long bottom set pin so it is impossible to set the roller cluster too far down), or you can wrap some gaffer's tape around the bottom part of the slidebar such that the roller cluster can't be set lower than the bottom deck's set pin.

 |  IP: Logged

Bruce McGee
Phenomenal Film Handler

Posts: 1776
From: Asheville, NC USA... Nowhere in Particular.
Registered: Aug 1999


 - posted 06-29-2008 10:36 PM      Profile for Bruce McGee   Email Bruce McGee   Send New Private Message       Edit/Delete Post 
Wow. I have trouble getting my people to even USE the bottom platters on our AW3's!

Glad you found out the cause of the scratching.

 |  IP: Logged

John Wilson
Film God

Posts: 5438
From: Sydney, Australia.
Registered: Dec 1999


 - posted 06-29-2008 11:01 PM      Profile for John Wilson   Email John Wilson   Send New Private Message       Edit/Delete Post 
I was taking up on the bottom plate (as I like to do) and the operator who was taking over from me said: "Why are you using the spare plate?"

[Roll Eyes]

 |  IP: Logged



All times are Central (GMT -6:00)  
   Close Topic    Move Topic    Delete Topic    next oldest topic   next newest topic
 - Printer-friendly view of this topic
Hop To:



Powered by Infopop Corporation
UBB.classicTM 6.3.1.2

The Film-Tech Forums are designed for various members related to the cinema industry to express their opinions, viewpoints and testimonials on various products, services and events based upon speculation, personal knowledge and factual information through use, therefore all views represented here allow no liability upon the publishers of this web site and the owners of said views assume no liability for any ill will resulting from these postings. The posts made here are for educational as well as entertainment purposes and as such anyone viewing this portion of the website must accept these views as statements of the author of that opinion and agrees to release the authors from any and all liability.

© 1999-2020 Film-Tech Cinema Systems, LLC. All rights reserved.