John Hawkinson
Film God

Posts: 2273
From: Cambridge, MA, USA
Registered: Feb 2002


 - posted 05-17-2008 12:58 AM      Profile for John Hawkinson   Email John Hawkinson   Send New Private Message       Edit/Delete Post 
Over in Best/worst projector, I asked people to justify their claims of projector steadiness with some quantitative numbers. No one bit [Frown]

So, on a pair of Century Cs, which couldn't possibly be all that great, here are some measurements of steadiness. I'm reminded that John Pytlak encouraged measuring steadiness not-from-a-loop, so here are some numbers for vertical steadiness from looking at perfs with the aperture pulled.

With this particular lens, the projected height of a perf was 19.75" -- a KS1870 perf is actually 0.078".

Projector #1: 5/16" * (078 mils/19.75") => 0.00123"
Projector #2: 7/16" * (078 mils/19.75") => 0.00173"

Or, in RP40 squares:

Projector #1: 0.00123/0.004125 => 0.299 squares
Projector #2: 0.00173/0.004125 => 0.419 squares

Turns out that's pretty bad. In this 1999 post, John Pytlak described the measurement standard used by Kodak's Screencheck:

quote:

Steadiness is measured using a roll of SMPTE 35-PA (RP40) test film. For vertical steadiness (jump), an "A" rating is less than 1/10 of a square movement, a "B" is 1/10 square, a "C" is 2/10, a "D" is 3/10 and an "F" is greater than 3/10. For horizontal movement (weave), "A" is less than 4/10 of a square movement, "B" is 4/10, "C" is 5/10, "D" is 6/10, and "F" is more than 6/10. These criteria are based on SMPTE Recommended Practice RP105, "Method of Determining the Degree of Jump and Weave...". As you know, each square on the 35-PA film represents 1/2 percent of image width, or 0.004125 inches on the film.

By that standard, my Projector #1 grades out at a D, and my #2 at an F. Pretty bad! [Embarrassed] Admittedly I think the gate tension was probably set waaay low, so I should probably reset it and rerun the test. But in the meantime, post your numbers!

--jhawk

 |  IP: Logged

John Walsh
Film God

Posts: 2490
From: Connecticut, USA, Earth, Milky Way
Registered: Oct 1999


 - posted 05-17-2008 10:48 AM      Profile for John Walsh   Email John Walsh   Send New Private Message       Edit/Delete Post 
Funny, I was just explaining to a friend about my doing the same test on some projectors 1982. I've tossed most of the notes but the one that remains lists of a pair of rebuilt JJ's, and one SA. I was using RP105-1981 (don't know if a newer practise has been published or, if so, if it's any different.) We couldn't get a long length of 35PA; (too expensive) we had to use 8ft loops.

The JJ's did OK at class 'B' (0.18%). The one SA at our 'dump' theater was a class 'D' (0.31%). Those are jump numbers; for some reason I didn't do weave. Not that it matters; none of those projectors are in service anymore. Also, few intermittent rebuilders are going to take an intermittent back and redo it.

 |  IP: Logged

Sean McKinnon
Phenomenal Film Handler

Posts: 1712
From: Peabody Massachusetts
Registered: Sep 2000


 - posted 05-17-2008 03:02 PM      Profile for Sean McKinnon   Author's Homepage   Email Sean McKinnon   Send New Private Message       Edit/Delete Post 
unfortunately, I no longer work for the company I used to work for where I serviced the projectors we were discussing.

 |  IP: Logged

Galen Murphy-Fahlgren
Master Film Handler

Posts: 405
From: Canton, MI, USA
Registered: Oct 2007


 - posted 05-18-2008 06:01 PM      Profile for Galen Murphy-Fahlgren   Email Galen Murphy-Fahlgren   Send New Private Message       Edit/Delete Post 
As I said, today was my first opportunity to do any sort of check on this front, and even today I only had time enough to eyeball it from booth, and only a couple of my projectors. All measures were made in scope. Here we go:

Aud. 7- Proj. V5 Horizontal movement was almost a full square. Vertical, about .5. F

Aud. 1 Proj. V5 Horizontal, approx. .75, vertical about .5 again. F

Aud. 4 Proj. Simplex XL Horizontal around .3 or .4. I could detect no vertical movement, steady as a rock. B?

Aud. 5 Proj. XL Horizontal .7 or so, vertical .4. F.

Again, these were eyeballed, and I'm lucky to have had time to do even that today. I will attempt to get some more accurate figures and figures for my other V5 and my pair of Christies. Also, I probably need a new RP40, I almost think mine is acetate. While that would obviously affect overall values because of gate tension being set for poly, could it have an effect on relative values, ie, some designs being better suited to accommodating different base thickness?

What steps can I take to improve the steadiness of these projectors? I mean, it isn't like I want the V5 to fail. I'm trying to provide good presentation, as I think most people are. I'm just doing it with very little experience and minimal guidance.

 |  IP: Logged

Sean McKinnon
Phenomenal Film Handler

Posts: 1712
From: Peabody Massachusetts
Registered: Sep 2000


 - posted 05-19-2008 09:58 AM      Profile for Sean McKinnon   Author's Homepage   Email Sean McKinnon   Send New Private Message       Edit/Delete Post 
I went through my files and pulled out my technician copies of my old service reports and PM reports. It seems like I recorded everything else except these measurements, I just noted either no or minimal horizontal movement (weave), or no or minimal vertical movement (jitter). Sorry, I tend to be pretty thorough but I guess not that thorough.

 |  IP: Logged

Richard P. May
Expert Film Handler

Posts: 243
From: Los Angeles, CA
Registered: Jan 2006


 - posted 05-19-2008 12:39 PM      Profile for Richard P. May   Email Richard P. May   Send New Private Message       Edit/Delete Post 
This is a good exercise. On our Norelco FP20s:
Proj 1 - C
Proj 2 - A

RPM

 |  IP: Logged

John Hawkinson
Film God

Posts: 2273
From: Cambridge, MA, USA
Registered: Feb 2002


 - posted 05-19-2008 01:01 PM      Profile for John Hawkinson   Email John Hawkinson   Send New Private Message       Edit/Delete Post 
Real numbers, Richard? I'll try to do ours more properly...

--jhawk

 |  IP: Logged

Steve Guttag
We forgot the crackers Gromit!!!

Posts: 12814
From: Annapolis, MD
Registered: Dec 1999


 - posted 05-19-2008 10:27 PM      Profile for Steve Guttag   Email Steve Guttag   Send New Private Message       Edit/Delete Post 
Today I looked at a Kinoton FP30E running 35PA...which is not always the most steady of test films...it varies batch-to-batch...

1/20th of a square jump (could have been closer to 1/30th...the screen was up there though I was trying to measure at the screen). So I guess that puts it in an "A" classification.

 |  IP: Logged

Michael Moore
Expert Film Handler

Posts: 188
From: Dover, DE / USA
Registered: Jun 2006


 - posted 05-21-2008 08:15 AM      Profile for Michael Moore   Author's Homepage   Email Michael Moore   Send New Private Message       Edit/Delete Post 
If i did this test with a screen say 10 feet away from the projector, (in our place I have room in the balcony to set up a smaller screen) will the test results be comparable to our 90 foot throw?

 |  IP: Logged

Paul Mayer
Oh get out of it Melvin, before it pulls you under!

Posts: 3836
From: Albuquerque, NM
Registered: Feb 2000


 - posted 05-21-2008 12:02 PM      Profile for Paul Mayer   Author's Homepage   Email Paul Mayer   Send New Private Message       Edit/Delete Post 
Since you're going to be measuring how much a given projected square moves in relation to its own height and width, the actual size of the square is not important. However, on such a small screen it may become difficult to precisely measure how much the square moves since it will be quite small.

 |  IP: Logged

Richard P. May
Expert Film Handler

Posts: 243
From: Los Angeles, CA
Registered: Jan 2006


 - posted 05-21-2008 02:15 PM      Profile for Richard P. May   Email Richard P. May   Send New Private Message       Edit/Delete Post 
This is a lab screening room, with about a 6 ft. high screen. An exact measurement is pretty hard to accomplish.
I re-viewed with a new, Estar based projection chart, originally created by the MGM Camera Dept. It's somewhat different from SMPTE, but with the necessary objects to compare the image with the screen edge or masking. With this, there is absolutely no discernable movement with either of our FP40s.

RPM

 |  IP: Logged

Carey Barber
Expert Film Handler

Posts: 143
From: Newport News, VA, USA
Registered: Jan 2003


 - posted 05-22-2008 11:57 AM      Profile for Carey Barber   Email Carey Barber   Send New Private Message       Edit/Delete Post 
What is the recommended length for a good test loop of 35PA?

 |  IP: Logged

Paul Mayer
Oh get out of it Melvin, before it pulls you under!

Posts: 3836
From: Albuquerque, NM
Registered: Feb 2000


 - posted 05-22-2008 01:30 PM      Profile for Paul Mayer   Author's Homepage   Email Paul Mayer   Send New Private Message       Edit/Delete Post 
For this kind of test a loop is not recommended, since a loop doesn't recreate the typical condition of film running through a projector. With each pass the loop heats up a bit more and changes the way the film behaves when it is stopped and flashed within the gate.

Loops are fine for doing things like checking and filing plates, but not for critical testing.

If you absolutely must use a loop, longer is better. Mine all happen to be 8-footers.

 |  IP: Logged

Carey Barber
Expert Film Handler

Posts: 143
From: Newport News, VA, USA
Registered: Jan 2003


 - posted 05-22-2008 09:10 PM      Profile for Carey Barber   Email Carey Barber   Send New Private Message       Edit/Delete Post 
Okay...This may be a stupid question, but what do you use in place of a loop?

Are you just using tons of 35PA and running it through? I would think that you would want a few minutes worth to give you time to check everything. That would be awfully expensive, as that test film is not cheap.

 |  IP: Logged

John Hawkinson
Film God

Posts: 2273
From: Cambridge, MA, USA
Registered: Feb 2002


 - posted 05-22-2008 09:18 PM      Profile for John Hawkinson   Email John Hawkinson   Send New Private Message       Edit/Delete Post 
I did my numbers from looking at perfs without the aperture plate.

If you run a reel of 35PA, you get much less wear than you do with a loop. It might not be economically reasonable for you, but it is the better way [Smile]

(We do use a 50' length of Cat69T for our Dolby Alignments, with heads and tails. That works pretty well, and lasts a long time...)

--jhawk

 |  IP: Logged



All times are Central (GMT -6:00)
This topic comprises 2 pages: 1  2 
 
   Close Topic    Move Topic    Delete Topic    next oldest topic   next newest topic
 - Printer-friendly view of this topic
Hop To:



Powered by Infopop Corporation
UBB.classicTM 6.3.1.2

The Film-Tech Forums are designed for various members related to the cinema industry to express their opinions, viewpoints and testimonials on various products, services and events based upon speculation, personal knowledge and factual information through use, therefore all views represented here allow no liability upon the publishers of this web site and the owners of said views assume no liability for any ill will resulting from these postings. The posts made here are for educational as well as entertainment purposes and as such anyone viewing this portion of the website must accept these views as statements of the author of that opinion and agrees to release the authors from any and all liability.

© 1999-2020 Film-Tech Cinema Systems, LLC. All rights reserved.